Posted January 22, 201510 yr We have a new Babylog 8000 plus ventilator in our NICU for the first time and I have 2 questions about it: 1- What is the commonly used values regarding volume triggering ? 2- what is the acceptable percentage of leak ? ــــــــــــــــــــــــــــــــــــــــــــــــــــــــــــــــــــــــــــــــــــــ Mohamad Ismail, Neonatology resident, Mansoura, Egypt .
January 26, 201510 yr 1.Volume guaranteed (VG) provides automatic adjustment of the peak inspiratory pressure for ensuring a minimum set tidal volume. The flow sensor of the ventilator should calibrated before starting VG ventilation. We set VT of 3.5-4 ml/kg for infants >1000 g and 5 ml/kg <1000 g. During SIMV/SIPPV/PSV + VG, all ventilator settings are the same, only the PIP limit sets 10 cm H2O above the PIP that is used during the conventional SIMV mode. At the follow-up, the set VT is adjusted according to the arterial blood gas analysis. 2. The leak is very important during VG ventilation, no more than 40%; higher leak results in erroneous measurement of delivered VTe (expire) and lead to a discrepancy between set VT and delivered VT. Best regards.
February 5, 201510 yr Author SirinDrMananParikh Thanks for your answers, but what about the degree of sensitivity that is commonly used ?